[Help] How to select the SAT temperature

[复制链接]

签到天数: 121 天

[LV.7]常住居民III

发表于 2018-6-7 09:28:15 | 显示全部楼层 |阅读模式
北京中仪天信科技有限公司
When I read the AMS-2750E, I found in this specification, it's absent for the temperature selection for a SAT.
For the calibration of Sensors, it stated that the temperature interval shall not greater than 140C within the measurement range.
For the calibration of instrument, it stated that at lease 3 point for equipment sensor, minimum, maximum and middle 1/3.
For the TUS, it stated that the temperature for survey shall be maximum, minimum and temperatures makes the difference between test temperatures not greater than 335C.
But for SAT, it stated no temperature for test?

do you guys have any comments on this?

签到天数: 121 天

[LV.7]常住居民III

 楼主| 发表于 2018-6-11 16:11:52 | 显示全部楼层
sorry for ignored the content in AMS-2750E.
The AMS-2750E actually stated the temperature requirement:
3.4.5.1 The displayed temperature indication and/or recording of the sensor being tested as used in production, with appropriate offsets or correction factors, at any operating temperature, shall be compared with the corrected temperature indication of the test sensor on a test instrument.

In this clause, it shows that it already gave the temperature requirement for SAT.

Thanks and for your information.
回复 支持 反对

使用道具 举报

签到天数: 247 天

[LV.8]以坛为家I

发表于 2018-6-25 17:55:51 | 显示全部楼层
I think you have got the answer by yourslf- at any operating temperature.
It makes no sense if not at the operating temperature.
回复 支持 反对

使用道具 举报

签到天数: 9 天

[LV.3]偶尔看看II

发表于 2018-10-17 14:47:10 | 显示全部楼层
Ha.
Usually, the people select one temperature he ofen used to perform SAT.
回复 支持 反对